[Gate-users] Re : SPECT intrinsic spatial resolution

Rajaram, Ramya (H USA) ramya.rajaram at siemens.com
Thu Sep 16 19:30:57 CEST 2010


Sorry about that - I mean a point source. Planar imaging I said to mean that the detector is not rotating or moving. The code for the source definition is as follows:

/gate/source/addSource SourceConfinement
/gate/source/SourceConfinement/gps/type Point
/gate/source/SourceConfinement/gps/centre 4. 0. 0. mm
/gate/source/SourceConfinement/gps/particle gamma
/gate/source/SourceConfinement/gps/energy 140. keV
/gate/source/SourceConfinement/setActivity 10.e-3 Ci
/gate/source/SourceConfinement/gps/angtype iso

-----Original Message-----
From: sebnem erturk [mailto:sebnem.erturk at gmail.com] 
Sent: Thursday, September 16, 2010 12:19 PM
To: Rajaram, Ramya (H USA)
Cc: gate-users at lists.opengatecollaboration.org
Subject: Re: [Gate-users] Re : SPECT intrinsic spatial resolution

what do you mean  by planar image of a point source?

what is the exact geometry of your source? (planar or point? )

cheers,

sebnem





On Thu, Sep 16, 2010 at 7:15 PM, Rajaram, Ramya (H USA) <ramya.rajaram at siemens.com> wrote:


	I am simulating planar image of a point source, so there is no phantom. Also, I have put in only the collimator and the continuous crystal detector - no back compartment or any other layers. I output the projection images using the the following code:
	
	/gate/output/projection/enable
	/gate/output/projection/setFileName file6
	/gate/output/projection/pixelSizeX 2.4 mm
	/gate/output/projection/pixelSizeY 2.4 mm
	/gate/output/projection/pixelNumberX 256
	/gate/output/projection/pixelNumberY 256
	
	Thanks,
	
	Ramya
	




	-----Original Message-----
	From: Samuel BURG [mailto:samuel_burg at yahoo.fr]
	Sent: Thursday, September 16, 2010 10:35 AM
	To: Rajaram, Ramya (H USA); gate-users at lists.opengatecollaboration.org
	Subject: Re : [Gate-users] SPECT intrinsic spatial resolution
	
	Yes, GATE do apply a spatial blurring in your projection when using
	/gate/digitizer/Singles/spblurring/setSpresolution
	
	but to help you, we need your full macro file.
	also remember that the collimator, phantom scattering and the background compartment of the PMT play a lot in sp resolution.
	
	 -----------------------------------------------------------
	Samuel Burg
	
	
	Mail : <samuel_burg at yahoo.fr>
	-----------------------------------------------------------
	"Aw, come on, who would be so stupid as to insert a cast
	to make an error go away without actually fixing the error?"
	Apparently everyone.
	-- Raymond Chen --
	-----------------------------------------------------------
	
	
	________________________________
	
	De : "Rajaram, Ramya (H USA)" <ramya.rajaram at siemens.com>
	À : gate-users at lists.opengatecollaboration.org
	Envoyé le : Jeu 16 septembre 2010, 16h 59min 48s
	Objet : [Gate-users] SPECT intrinsic spatial resolution
	
	Dear Gaters,
	
	Does GATE really convolve the intrinsic spatial resolution for SPECT? I
	have generated 256x256 projection images from two macros, both identical
	except for the fact that in one, the spResolution is set to 3.7 mm, in
	the other it is set to 0 mm and yet, the two images look identical. A
	simple profile also gives me the same system resolution for both - leads
	me to think the spResolution is somehow not being convolved -- could
	someone please explain/help?
	
	The following code is what I used to set the digitizer up:
	
	/gate/digitizer/Singles/insert adder
	/gate/digitizer/Singles/insert blurring
	/gate/digitizer/Singles/blurring/setResolution 0.09
	/gate/digitizer/Singles/blurring/setEnergyOfReference 140. keV
	/gate/digitizer/Singles/insert spblurring
	/gate/digitizer/Singles/spblurring/setSpresolution 0. mm (3.7 mm)
	/gate/digitizer/Singles/spblurring/verbose 2
	/gate/digitizer/Singles/insert thresholder
	/gate/digitizer/Singles/thresholder/setThreshold 50. keV
	/gate/digitizer/Singles/insert upholder
	/gate/digitizer/Singles/upholder/setUphold 200. keV
	
	Thank you very much,
	
	Ramya
	
	----------------------------------------------------------------------------
	This message and any included attachments are from Siemens Medical Solutions
	and are intended only for the addressee(s).
	The information contained herein may include trade secrets or privileged or
	otherwise confidential information. Unauthorized review, forwarding, printing,
	copying, distributing, or using such information is strictly prohibited and may
	be unlawful. If you received this message in error, or have reason to believe
	you are not authorized to receive it, please promptly delete this message and
	notify the sender by e-mail with a copy to Central.SecurityOffice at siemens.com
	
	Thank you
	_______________________________________________
	Gate-users mailing list
	Gate-users at lists.opengatecollaboration.org
	http://lists.opengatecollaboration.org/mailman/listinfo/gate-users
	
	
	_______________________________________________
	Gate-users mailing list
	Gate-users at lists.opengatecollaboration.org
	http://lists.opengatecollaboration.org/mailman/listinfo/gate-users
	




-- 