Several weeks ago, I reported an phenomenon about the strange random
coincidence distribution among the detectors with a point source placed
in the center of the detectors. Recently I found out why it happened.
The problem is caused by the sequence of the deadtime and the energy-cut
macros in the
digitizer segment. In GATE Users Guide and all the benchmark or demo
files, the
definitions of deadtime are always placed after the energy-cut. However, I
think this might not be the real case. We know one detector should be dead
for a certain time (deadtime) after receiving one event, no matter how
much energy has
been deposited by this event. So even the events with a very small or very
large energy should
make the detector dead for a while. But if the deadtime is defined after
the
energy-cut, those events with an energy that is outside the energy window
will be cut
and can not reach the module of deadtime in the digitizer chain. Therefore
those
detectors can still detect another event immediately. In this case, the
coincidences rates
will become larger than the real one.
Best regards,
Yuxuan Zhang
Dr. Yuxuan ZHANG
Dept. Experimental Diagnostic Imaging
Univ. Texas, MD Anderson Cancer Center
1515 Holcombe Blvd, Unit 217
Houston, TX 77030-4095
Tel: +1-713-745-1671
Fax: +1-713-745-1672
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://lists.opengatecollaboration.org/mailman/private/gate-users/attachments/20040927/c95f8dd4/attachment.htm>
More information about the Gate-users
mailing list