[Gate-users] Is the time resolution for singles?
Martín Belzunce
martin.sure at gmail.com
Wed Jun 9 16:32:12 CEST 2010
Hi Jinxia,
I think you are right, the time resolution is set for the singles. In
the documentation this module is applied to the singles chain, so it
should be that way. A few months ago, I had the same doubt so I ran a
simulation to try this module. I used a digitizer with a time
resolution module of 4.13 nseg FWHM. This means that the FWHM for the
coincidences should be 4.13*sqrt(2) = 5.84 nseg. I processed the
simulation and made an histogram of time difference between the time
stamps of the two events of each coincidences, and I got the plot I
attach here. The FWHM was 5.9 nseg. With these results I could verify
that time resolution was set to the singles.
Best Regards,
Martín Belzunce
On Thu, Jun 3, 2010 at 11:11 AM, Jinxia Guo <jinxia.guo2009 at gmail.com> wrote:
> Hello users,
>
> In GATE simulation, is the time resolution for singles? For example, if you
> set the time resolution to 900ps, it means that the FWHM of the singles'
> time blurring is 900ps, not 900/sqrt(2)ps. is it? Thank you for help!
>
> Regards,
> Jinxia
>
>
> --
> Jinxia Guo
> School of Life Sci&Tech
> D11 Building Room 503
>
> Huazhong Univ. of Sci. & Tech
>
> 1037 Luoyu Road, Wuhan 430074, China
> Tel:+86-27-8779-2294
> _______________________________________________
> Gate-users mailing list
> Gate-users at lists.opengatecollaboration.org
> http://lists.opengatecollaboration.org/mailman/listinfo/gate-users
>