[Gate-users] Re: Difference in energy spectrum for first and second event in cylindrical PET system
Paul Vaska
vaska at bnl.gov
Wed Aug 29 18:05:36 CEST 2007
Perhaps Yuxuan and I are both correct about the ordering of events in
the coincidence output. According to Yuxuan, for true coincidences,
the higher energy single is listed first (single 1). That is why
there is no fixed time order for true events and why the energy
spectra for single 1 and single 2 differ. On the other hand, for
random coincidences, the event that is earlier in time is listed
first. Note that the coincidence output corresponds to the prompt
time window and thus includes both true and random coincidences (and
the energy spectra for single 1 and single 2 will be a mixture). For
the delayed coincidence output, because the order depends on event
time, I agree with Andrew that the energy spectra should not differ.
However, this may be an artificial problem. My point about sorting
by detector or detector pair is so that you avoid binning your data
based on artificial properties such as the order that events are
written to an output file. What really matters is the energy
spectrum for a given detector, not the spectrum for a given column in
an output file. Therefore, I'm not too worried that single 1 and
single 2 have different spectra (because 1 and 2 do not correspond to
any physical property), but would be very concerned if detector a and
detector b had different spectra (because this is not physically
possible).
Paul
Paul Vaska, Ph.D.
Medical Department, Building 490
Brookhaven National Laboratory
Upton, NY 11973
vaska at bnl.gov (631)344-6228 fax:(631)344-4146
On Aug 28, 2007, at 6:44 PM, Andrew Goertzen wrote:
> Hi Paul and Yuxuan,
>
> Following your suggestions, I took a closer look at the energy
> histogram
> of the delayed coincidences by looking at the delays.dat output file.
> As Paul said, the order of the singles in a random coincidence is
> always
> that the early single is the first half and single 2 (the later
> one) is
> always the second half. In the delays file, the two singles should
> be,
> by definition, just singles and thus they should exhibit an identical
> energy spectrum. This is not what I see when I look at the data (see
> attached energy histogram). In my data, single 1 has a different
> energy
> spectrum than single 2. The only real difference between the two
> singles is that single 1 is 1 microsecond earlier than single 2. I
> have
> no idea why the two singles in the delays file should have different
> spectra, so any additional insight you could offer into this would be
> appreciated.
>
> Cheers,
> Andrew
>
> Andrew Goertzen
> Department of Radiology
> University of Manitoba
>
>
>
>
> <DelaysEnergyHistogram.jpg>
> This email and/or any documents in this transmission is intended
> for the
> addressee(s) only and may contain legally privileged or
> confidential information. Any unauthorized use, disclosure,
> distribution, copying or dissemination is strictly prohibited. If
> you receive this transmission in error, please notify the sender
> immediately and return the original.
>
> Ce courriel et tout document dans cette transmission est destiné à
> la personne ou aux personnes à qui il est adressé. Il peut contenir
> des informations privilégiées ou confidentielles. Toute
> utilisation, divulgation, distribution, copie, ou diffusion non
> autorisée est strictement défendue. Si vous n'êtes pas le
> destinataire de ce message, veuillez en informer l'expéditeur
> immédiatement et lui remettre l'original.
More information about the Gate-users
mailing list